P1: A temperature measuring system is exited by a 0.25 Hz harmonically varying input. If the time constant of the system is 4.0s and the indicated amplitude is 10 degrees celcius, what is the true temperature?
P2: A thermocouple with a time constant of 0.05s is considered to behave as a first order system, over what frequency range can the thermocouple measure dynamic temperature fluctuation (which is harmonic) with an error less than 5%?
P3: A force transducer behaves as a second-order system. If the undamped natural frequency is 1800 Hz and its damping is 30% of the critical, (a) determine the error in the measured force for a harmonic input of 950 Hz. (b) What is the magnitude in the phase angle?




P1: A temperature measuring system is exited by a 0.25 Hz harmonically varying input. If the...
QUESTION 4 (20 marks) For any given experiment data of a sensor, explain how to determine a. to ench 63.2 2 ofHe Panes the settling time of a second-order system Tire to at 22 Tire the time constant of a first-order system (i) () b. A thermocouple with a time constant of 0.05s is considered to behave as a first-order system. For the thermocouple to measure the dynamic temperature with an error less than 5%, what is (8) the range...