Many medical devices use a footswitch to toggle important functions (like turning a laser ON). The...
Many medical devices use a footswitch to toggle important functions (like turning a laser ON). The footswitch is affixed in metal housing to prevent accidental use and possible damage. See the figure. The housing is required to withstand at least 300 lb force before failure. A failure is a complete collapse of the housing such that the switch is affected 3. To test the housing strength a 20 kg mass is dropped onto the housing from a height of 2 m. We assume that the 20 kg mass slows down linearly upon impact. At impact, the top of the housing buckles 5.5 cm. From energy considerations, calculate the impact force needed to buckle the housing a. b. Does the housing meet requirements?