Question

If two bands are not fully resolved in a reverse-phase liquid chromatography experiment when using a...

If two bands are not fully resolved in a reverse-phase liquid chromatography experiment when using a mobile phase of 50:50 water: acetonitrile, which of the two components of the mobile phase should you increase the percentage of in order to improve the resolution? Explain how this change will improve the peak separation?

0 0
Add a comment Improve this question Transcribed image text
Answer #1

In a reverse phase chromatography stationary phase is nonpolar and mobile phase is Polar

so as mobile phase moves up it separates the non polar compound from polar. it takes polar molecules with it, thereby achieving a separation.

in this case given in the question to improve the resolution we need to to increase polarity of the solvent. for this the amount of acetonitrile in mobile phase solvent should be increased because acetonitrile is more polar than water.

This will give a better and good resolution. This will result in better movement of Polar molecules thereby giving a good peak separation.

Add a comment
Know the answer?
Add Answer to:
If two bands are not fully resolved in a reverse-phase liquid chromatography experiment when using a...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
  • 5. Liquid Chromatography. b. For a reverse-phase HPLC separation with a mobile phase pH = 3.00,...

    5. Liquid Chromatography. b. For a reverse-phase HPLC separation with a mobile phase pH = 3.00, circle which of the following compounds would elute first and briefly describe why: Propanoic acid (CH3CH2COOH, PK. = 4.87) or Propylamine (CH2CH2CH2NH2, pks = 3.43). b. If one uses gradient elution with normal phase chromatography, would it be better to go from a less polar to a more polar solvent or a more polar to a less polar solvent? Why?

  • PART 1: In a gas chromatography experiment using a non-polar stationary phase A) solutes are eluted...

    PART 1: In a gas chromatography experiment using a non-polar stationary phase A) solutes are eluted in order of increasing boiling point. B) solutes are eluted in order of increasing vapor pressure. C) solutes are eluted in order of increasing polarity. PART 2: You can improve the separation of components in gas chromatography by A) raising the temperature of the injection port. B)raising the temperature of the column isothermally. C) raising the temperature of the column programmatically.

  • question #2 please Reverse phase chromatography is the most common type of liquid chromatography which the...

    question #2 please Reverse phase chromatography is the most common type of liquid chromatography which the solid phase is made up of a packing material that is largely hydrophobic and the eluents are of varying polarities. In general, the more polar compounds elute first from the column followed by more hydrophobic compounds. PRELAB QUESTIONS 1. Draw structures for the following analytes for the simulation: a. acetophenone b. benzophenone c. butylparaben d. propiopheneone e. ethylparaben f. propylparaben g. ketoprofen h. 3-nitrophenol...

  • 7. Given a non-polar stationary phase C18 (reverse phase) and a slightly polar mobile phase (15%...

    7. Given a non-polar stationary phase C18 (reverse phase) and a slightly polar mobile phase (15% water 85% acetonitrile) predict the order of elution for the following compounds (with 1 being the fastest and 4 being the slowest) (6 pts) Methanol CH OH 1-florobutane CH CH CH2CH2F Butanol CH,CH CHCHOH Cyclohexane CoH12 Briefly justify how you made your choice(s) Now increase the polarity of the mobile phase to 85% water and 15 % acetonitrile. Explain the relative shifts in retention...

  • 7. (a) When you try separating an unknown mixture by reversed-phase chromatography with 50% acetonitrile/50% water,...

    7. (a) When you try separating an unknown mixture by reversed-phase chromatography with 50% acetonitrile/50% water, the peaks are too close together and are eluted in the range k = 2-6. Should you use a higher or lower concentration of acetonitrile in the next run? (b) When you try separating an unknown mixture by normal-phase chromatography with 50% hexane/50% methyl t-butyl ether the peaks are too close together and eluted in the range k = 2-6. Should you use a...

  • 3. There are two major components in the chromatography experiment: the solvent system (mobile phase) and...

    3. There are two major components in the chromatography experiment: the solvent system (mobile phase) and the paper (stationary phase). Which of these major components is more polar phase and which is less polar phase? Explain using Lewis structures. (please PLEASE use the lewis structures, thats the part of the question I don't have yet!) 4. What parts of the molecules in the stationary phase will interact with the molecules that are being separated? Please answer this question using the...

  • Over the counter cold medications are complex mixtures that contain a combination of active ingredients that...

    Over the counter cold medications are complex mixtures that contain a combination of active ingredients that have specific effects. These mixtures are routinely analysed via high performance liquid chromatography (HPLC) to verify that the concentration of active ingredients are consistent to guidelines A cough mixture containing phenylephrine hydrochloride (decongestant) and chlorpheneramine maleate (antihistamine) was run on a HPLC with the parameters shown below Gemini C18 (150 x 4.6 mm, 5 ?m) HPLC column 1.0 mL/min UV-VIS at 215nm Column Flow...

  • 1. Select an appropriate type of column for separating these steroids and give the rationale for...

    1. Select an appropriate type of column for separating these steroids and give the rationale for your selection. HO HO ОН OH Prednisolone Prednisone 2. You are using a C18 Reverse Phase column and the parameters shown below to separate Acetaminophen, Caffeine, and Benzoic Acid. You observe three peaks in your chromatogram (Huzzah!). Which corresponds to which compound? Explain the rationale behind your assignment. OH C18 REVERSED PHASE COLUMN Acetaminophen HC N H, O Caffeine Column: Legacy L1 Size: 4.6...

  • Background In this lab, we are going to learn how to use a powerful separation technique...

    Background In this lab, we are going to learn how to use a powerful separation technique called chromatography which can separate compounds that are quite similar to one another. Chromatography is a general type of separation in which a mixture of compounds passes through a stationary phase. Different compounds have different relative affinities for the mixture traveling along the stationary phase, or the mobile phase, and the stationary phase, causing these compounds to separate from one another. More specifically, we’re...

  • Table 2: Alcohol and Ketone Standards GC Retention Times GC#: Retention time Alcohol Standard Mixture Phent...

    Table 2: Alcohol and Ketone Standards GC Retention Times GC#: Retention time Alcohol Standard Mixture Phent ion= 2.05Smin low Peak 1 Compound Name: athanol Rapesnal utanal aclapeatanal 4.755min oilng 6.895min .230 min Peak 2 Compound Name: Peak 3 Compound Name: Peak 4 Compound Name: GC#:1 Ketone Standard Mixture Retention time 2.575un Peak 1 Compound Name: Acclono Butanon a Pentanone 2tHex anon Peak 2 Compound Name: min Peak 3 Compound Name: 615 min 1D. 307 min Peak 4 Compound Name: Part...