When an ion chamber is used for reference dosimetry measurements, what are the required corrections to convert electrometer reading into dose? Why we use each of them.

0 0
Add a comment Improve this question Transcribed image text
Answer #1

Electrometer calibration (Kelec ) Values An electrometer calibration correction factor is employed when the ionization chamber and the electrometer are calibrated separately. (If they are calibrated together, (Kelec) Q is not required.) The calibration certificate states that (Kelec) Q = 1.000 ± 0.25%. Uncertainties The estimated standard uncertainty is 0.14%, assuming a rectangular distribution with the variation limits set at ±0.25%.

Corrected reading M*Q

The components that we took into account were • charge reading, • monitor chamber, • pressure, • temperature, • relative humidity, • electrometer calibration, • polarity, and • saturation.

Charge reading

Values MQ = MDisplay − MLeakage ≅ MDisplay

The value of MLeakage can be determined by measuring the collected charge in the absence of radiation after and before the irradiation. That reading can then be used to verify that the reading from the electrometer is MDisplay ≥ 103 MLeakage, as recommended in TRS-398, and so on, accomplishing condition.

Uncertainties

(MQ) = u2 (MDisplay) + u2 (MLeakage)

Air kerma based protocol:

Air kerma based protocols use the air kerma in air calibration coefficient NK, Co-60 obtained for a local reference ionization chamber in a Co-60 beam at a standards laboratory. Routine ionization chambers are then cross-calibrated with the reference ionization chamber in a local Co-60 beam. Two steps are involved in an air kerma based protocol for the calibration of megavoltage photon beam

A formalism based on an air kerma calibration of the user, ionization chamber is recommended. It is recommended that the transfer be carried out according to the procedure of calibration The cavity air calibration coefficient N Dair is defined as: air,Co ND.air = MCo (3.31) Where Daue ts the alsored dose to air in the chamber cavity. Mco its the chamber signal corrected for influence quantities Kairair as follows: D1-g)K*K 8.32) Where g is the radiative fraction, i.e., the fraction of the total transferred energy expended in radiative interactions on the slowing down of the secondary electrons in airAbsorbed dose to water based protocol:

All dosimetry protocols aim at determination of the quantity absorbed dose to water. It is therefore logical to provide ionization chambers directly with a calibration coefficient in terms of this quantity, rather than in terms of the air kerma in air, if at all possible. Recent developments have provided support for a change in the quantity used at present to calibrate ionization chambers and provide calibration coefficients in terms of absorbed dose to water ND,W for use in radiotherapy beams. Many PSDLs now provide ND,W


answered by: ANURANJAN SARSAM
Add a comment
Know the answer?
Add Answer to:
When an ion chamber is used for reference dosimetry measurements, what are the required corrections to...
Your Answer:

Post as a guest

Your Name:

What's your source?

Earn Coins

Coins can be redeemed for fabulous gifts.

Not the answer you're looking for? Ask your own homework help question. Our experts will answer your question WITHIN MINUTES for Free.
Similar Homework Help Questions
ADVERTISEMENT
Free Homework Help App
Download From Google Play
Scan Your Homework
to Get Instant Free Answers
Need Online Homework Help?
Ask a Question
Get Answers For Free
Most questions answered within 3 hours.
ADVERTISEMENT
ADVERTISEMENT