Low concentrations of warfarin near the detection limit gave the dimensionless instrument readings: 203.3, 172.3, 160.3,...
Help!!! :(
Low concentrations of thallium near the detection limit gave the following dimensionless instrument readings: 206.9,173.7.163.9,175.9, 220.9.161.7, 224.7, 203.3,136.3, and 207.1. Ten blanks had a mean reading of 42.1. The slope of the calibration curve is 1.76 times 10^9M^-1. Estimate the signal and concentration detection limits and the lower limit of quantitation for thallium. Signal Detection Limit: Concentration Detection Limit: Lower Limit of Quantitation:
A researcher is interested in determining the detection and quantitation limits of a method for the detection of warfarin. Ten method blanks gave the dimensionless instrument readings: 72.3 , 34.1 , 29.3 , 41.3 , 86.3 , 27.1 , 79.1 , 64.7 , 31.7 , and 72.5 . Ten samples containing a low concentration of warfarin near the detection limit had a mean reading of 189.1 . The slope of the calibration curve is 2.50×109 M−1 . Estimate the signal...
Find the detection limit for a low concentration of analyte in mu M (to 2 significant figures), if the reagent blanks gave values of a detector signal of 1.4, 2.2, 1.7, 0.9, 0.4, 1.5, and 0.7 nA, and the slope of the calibration curve was m = 0.229 nA/ mu M.