A researcher is interested in determining the detection and quantitation limits of a method for the detection of warfarin. Ten method blanks gave the dimensionless instrument readings: 72.3 , 34.1 , 29.3 , 41.3 , 86.3 , 27.1 , 79.1 , 64.7 , 31.7 , and 72.5 . Ten samples containing a low concentration of warfarin near the detection limit had a mean reading of 189.1 . The slope of the calibration curve is 2.50×109 M−1 . Estimate the signal and concentration detection limits and the lower limit of quantitation for warfarin.
by using all number calculate mean and standard deviation by using formulas
mean = 53.84 = X
standard deviation = 23.22 = S
signal detection limit = Yblank + 3 S
= 189.1 + 3 x 23.22
= 258.8
signal detection limit = 258.8
concentration detection limit = 3S / M
= 3 x 23.22 / 2.50 x 10^9
= 2.79 x 10^-8 M
concentration detection limit = 2.79 x 10^-8 M
lower limit of qunatitation = 10 S / M
= 10 x 23.22 / 2.50 x 10^9
= 9.29 x 10^-8 M
lower limit of qunatitation = 9.29 x 10^-8 M
A researcher is interested in determining the detection and quantitation limits of a method for the...
Low concentrations of warfarin near the detection limit gave the dimensionless instrument readings: 203.3, 172.3, 160.3, 172.3, 217.3, 158.1, 221.1. 199.7, 132.7, and 203.5. Ten blanks had a mean reading of 56.3. The slope of the calibration curve is 3.55 x 10' M-!. Estimate the signal and concentration detection limits and the lower limit of quantitation for warfarin. signal detection limit: concentration detection limit: lower limit of quantitation:
Help!!! :(
Low concentrations of thallium near the detection limit gave the following dimensionless instrument readings: 206.9,173.7.163.9,175.9, 220.9.161.7, 224.7, 203.3,136.3, and 207.1. Ten blanks had a mean reading of 42.1. The slope of the calibration curve is 1.76 times 10^9M^-1. Estimate the signal and concentration detection limits and the lower limit of quantitation for thallium. Signal Detection Limit: Concentration Detection Limit: Lower Limit of Quantitation: