Q5: (2) DNA quantitation. You measure the absorption spectrum for a sample of DNA and obtain the values listed below. The DNA had been diluted 10 fold. Remember that a solution of DNA that is 50μg/ml (or 50ng/μl) has an absorbance of 1.0 at a wavelength of 260 nm. What is the concentration of the DNA in this sample (In this question DO NOT ignore the problem of dust in the sample and give answer in ng/ul).
A260 = 0.80 A280 = 0.40 A320 = .05 A230 = 0.40
Nanodrop takes the measurement at 260nm.
The ratio of 260/280 gives us details about the DNA quality.
If the ratio is between around 1.8, DNA is considered pure. Below 1.8 ratio DNA has the contamination of protein and phenol.
Given that A260 = 0.80 and A280 = 0.40
Ratio of 260/280 is 2. So DNA is not pure.
Absornace at 260 is 0.8.
So the concentration of DNA is 0.8*50 = 40ng/ul,
DNA was diluted 10 times before taking the reading on Nanodrop.
So DNA concentration in the original DNA sample is 10*40 = 400ng/ul.
Q5: (2) DNA quantitation. You measure the absorption spectrum for a sample of DNA and obtain...
DNA quantitation. You measure the absorption spectrum for a sample of DNA and obtain the values listed below. The DNA had been diluted 10 fold. Remember that a solution of DNA that is 50μg/ml (or 50ng/μl) has an absorbance of 1.0 at a wavelength of 260 nm. What is the concentration of the DNA in this sample (In this question DO NOT ignore the problem of dust in the sample and give answer in ng/ul). A260 = 0.80 A280...
After purification of your favorite plasmid you use a
spectrophotometer to find the concentration of your DNA and observe
the graph below. Resketch by hand this graph on your answer page
with exact axis labels and scales. Redraw the red line. Draw, in
your estimation, the absorption spectrum of what is expected of a
good DNA prep.
OD o X Load your sample and press the measure button. Blank Re-Blank Sample ID: Pedestal ort pectra ple volume Type: DNA 50.00...